1. Zero Adjustment in Series Ohmmeter: "Zero adjustment" refers to the process of making the needle point to "0 ohms" (full-scale deflection) when the test leads are shorted together.
2. Role of the Shunt Resistance ($R_2$): In a standard series ohmmeter, a variable resistor ($R_2$) is connected in shunt (parallel) with the meter movement.
• As the battery voltage decreases over time, the total current in the circuit drops.
• By adjusting the shunt resistor $R_2$, we change the proportion of current diverted through the meter.
• Increasing $R_2$ forces more current through the meter, allowing it to reach full-scale deflection (0 ohms) even with a slightly weaker battery.
3. Why not the series resistor?: While changing the main series resistance ($R_1$) could technically change the current, it would also significantly alter the "mid-scale" calibration of the meter, making the non-linear scale inaccurate across the rest of the range. Shunt adjustment is the standard method for compensation.