Question:medium

An analog instrument has a specified accuracy of \(\pm 1%\) of full-scale reading. If its full-scale value is 300 V and it reads 120 V, the maximum percentage error with respect to the indicated reading is

Show Hint

The absolute error remains constant at 3 V across the entire scale. As the measured voltage reading decreases, this fixed 3 V uncertainty becomes a larger percentage of the reading: \(\frac{3}{120} \times 100 = 2.5%\).
Updated On: Jun 25, 2026
  • \(1%\)
  • \(1.5%\)
  • \(2.5%\)
  • \(5%\)
Show Solution

The Correct Option is C

Solution and Explanation